Hi, i did not get a calibration certificate with my P8541 unit.
I need to get the unit calibrated urgently.
What is the procedure to calibrate this unit, because it is a digital probe?
TSensor has calibrate tab, but it is greyed out.
Offline
Hi,
P8541/P8511/P8631 devices use digital probes. Measurement accuracy depends on probe accuracy only. The device does not affect to accuracy.
If you buy P85x1 separately without probes calibration certificate is not created. If you buy device together with probes calibration certificate is attached. Calibration certificate to device without probes not make sense.
Calibration can be made according Calibration manual for Txxxx and Hxxxx attached with TSesnor software. Important is side 1 and 2 of this manual. Actual measured values can be readed using TSensor software or from web server. Adjustment with P85xx/P86xx devices is not possible because digital probe/sensor are used.
John